I was able to find accelerated indoor aging data for Flexible 80A V2 according to ASTM D4459, which simulates polymer aging due to solar radiation transmitted through glass. After 1,000 hours, the material retained approximately 87% of its initial elongation at break, indicating good resistance to this type of aging.
For the actual application, ASTM D4329 would be more relevant because it exposes plastics to fluorescent UVA, elevated temperature, and moisture/condensation. Mechanical properties are then measured after exposure as per ASTM 638.
Formlabs performed this test on Tough 1500 V2, where elongation at break dropped from approximately 155–160% initially to around 80–85% after 1,000 hours, while tensile strength remained approximately unchanged.
Since Formlabs has not published ASTM D4329 data for Flexible 80A V2, I also looked at comparable flexible/elastomeric vat-photopolymer materials with published aging data:
• 3D Systems Figure 4 RUBBER-65A — approximately 92% of EAB retained under ASTM D4329, with the other mechanical properties remaining relatively stable.
• 3D Systems Figure 4 RUBBER-BLK 10 — approximately 92% of EAB retained under ASTM D4329, with some changes in other mechanical properties.
• BASF / Forward AM Ultracur3D FL 60 — tested under ISO 4892-2 Method A rather than ASTM D4329. After 1,000 hours of UV exposure, EAB changed from 78% to 75% (~96% retention), while tensile strength changed from 3.0 MPa to 3.2 MPa.
Combined with the heat-aging results reported for Flexible 80A V2 under ASTM D3045 at 50 °C for six weeks, the available data suggest that Flexible 80A V2 has reasonably good environmental-aging resistance.
Therefore, although there is no ASTM D4329 result available specifically for Flexible 80A V2, based on its D4459 aging performance, heat-aging performance, and the D4329/UV-aging behavior of comparable flexible photopolymers, I would expect the part to survive the specified conditions with some reduction in elongation at break.
However, an exact property-retention value cannot be stated without testing Flexible 80A V2 itself under ASTM D4329.